  • 1.  New SIMOA NFL v2 kit and partial validation

    This message was posted by a user wishing to remain anonymous
    Posted 12-01-2022 10:48
    This message was posted by a user wishing to remain anonymous

    Dear members,
    Quanterix released a new version of their Nf-L kit called v2. In this new version, they have implemented a new type of beads but all the rest remain unchanged.

    We have just completed the validation using version 1 of the kit but this kit is no longer available. We are considering a partial validation but are wondering about the extent of it. I would really appreciate your input.

    Would 3 P&A runs be sufficient? Do we have to compare the 2 kits or only use v2 kits for this partial validation has no data from clinical samples has been yet generated?

    Thank you.

    Kind regards.


  • 2.  RE: New SIMOA NFL v2 kit and partial validation

    Posted 12-05-2022 11:52
    Hello,

    Thank you for submitting this question to the group. Are the previous version kits available to conduct a comparison between both versions of the kit? I would tend to think that 3 P&A runs would be appropriate for a partial validation, to include endogenous QCs in those A&P runs. If the previous kit version is not available, you can at least compare to historical data from the previous kit. 

    I'm also adding in @Yan Ni for experience with Nf-L.

    Best,
    Amanda​

    ------------------------------
    Amanda Hays Ph.D.
    Scientific Officer
    BioAgilytix
    Olathe KS
    [email protected]

    Disclaimer: Opinions expressed are solely my own and do not express the views or opinions of my employer.
    ------------------------------



  • 3.  RE: New SIMOA NFL v2 kit and partial validation

    This message was posted by a user wishing to remain anonymous
    Posted 12-06-2022 07:12
    This message was posted by a user wishing to remain anonymous

    Hello,
    thank you for your input.

    The 2 versions are still available. However, we used the V1 kit only for validation and not yet for clinical sample analysis. 

    Therefore, I thought that it would not be necessary to compare the 2 versions of the kits and that only showing that the new version of the kit has good performances could be sufficient.

    Kind regards.




  • 4.  RE: New SIMOA NFL v2 kit and partial validation

    Community Leadership
    Posted 12-12-2022 14:13
    Hello, sorry for the late reply, I have just returned from PTO..  
    I agree bridging is not required if data has not been generated with the version 1 kit.  In term of partial validation, in addition to P&A runs, you may also inquire whether the beads change affecting kit/reagent stability.  @Karen Quadrini @[email protected] have experience with this NfL kit change.  Looping them in for comment.
    Kind regards, 
    Yan
     ​​

    ------------------------------
    Yan Ni
    Executive Director
    Passage Bio
    Philadelphia PA
    [email protected]

    Disclaimer: Opinions expressed are solely my own and do not express the views or opinions of my employer.
    ------------------------------



  • 5.  RE: New SIMOA NFL v2 kit and partial validation

    Community Leadership
    Posted 12-21-2022 22:35
    Hello, apologies for the delay in response.  The kit has undergone changes in the assay diluent (detergent). In addition to Yan's and Amanda's suggestions, consider a parallelism assessment to confirm MRD and dilution range for the matrix to be used.  @Seema Gupta may have additional insight as well. ​

    ------------------------------
    Karen Quadrini, PhD
    Associate Director, Biomarkers and Precision Medicine
    Passage Bio
    Philadelphia, PA
    [email protected]

    Disclaimer: Opinions expressed are solely my own and do not express the views or opinions of my employer.
    ------------------------------



  • 6.  RE: New SIMOA NFL v2 kit and partial validation

    Posted 01-03-2023 07:13
    Good Morning, 
    very interesting suggestions.

    It seems that depending if the kit is run on HD-X or SR-x the buffer may not be different as the version 1 kit.