To define drug tolerance, I believe most people following the recommendation (or a variation) of the 'Report on the AAPS Immunogenicity Guidance Forum' (https://doi.org/10.1208/s12248-019-0328-8) which states: Concentrations of drug selected for evaluation should target concentrations that are at lower-than-expected, at expected, and at higher-than-expected levels for sample trough. PC should be tested at the LPC concentrations, 100, 250, and 500 ng/mL, and potentially higher concentrations. Drug tolerance should be evaluated in both the screening and confirmatory tiers.
Of the concentration tested, the 100, 250 and 500 ng/mL of PC are the most relevant for defining drug tolerance. Drug tolerance at the LPC is pretty irrelevant - the LPC is typically selected to be very close to the screening cut-point to monitor/control that there is consistent sensitivity between different runs. By this selection process, the LPC is often intolerant to even low drug levels.
Even though drug tolerance is defined relative to a 'surrogate' positive control, if the PC is a rabbit polyclonal, then I think we have enough experience in the field to say that if you can detect a positive response from 100 or 250 ng/mL PC in the presence of Ctrough levels of drug, then you're very likely to detect any ADA that are clinically relevant.
Best wishes,
Rob
------------------------------
Robert Nelson
Scientific Officer, Senior Director
BioAgilytix Europe GmbH
[email protected]Disclaimer: Opinions expressed are solely my own and do not express the views or opinions of my employer or other entities to which I am affiliated.
------------------------------
Original Message:
Sent: 05-29-2025 17:34
From: Anna Schwartz
Subject: Drug Tolerance definition for ADA result interpretation
Greetings! I would like to ask how do community members define Drug Tolerance (DT) for ADA methods. For example, in order to judge a sample that tested Negative in the screening assay as ADA-Negative, the concentration of the drug present in the sample should be at or below the method's DT level. Is everyone still using DT value corresponding to 100 ng/mL of Positive Control for this purpose, or do you use DT value corresponding to the "true" LPC concentration that was statistically calculated based on the method's sensitivity? Thank you!
------------------------------
Anna Schwartz
Associate Director, Bioanalysis
Kyowa Kirin, Inc.
Princeton NJ
[email protected]
------------------------------